    Excessive variations of K meter readings

    UIFlow
    1
    2
    1.6k
    Loading More Posts
    • Oldest to Newest
    • Newest to Oldest
    • Most Votes
    Reply
    • Reply as topic
    Log in to reply
    This topic has been deleted. Only users with topic management privileges can see it.
    • C
      csierra67
      last edited by

      Hi there,

      I am experiencing with the thermocouple K unit.
      The unit shows ambiant temperature and thermocouple temperature.
      I am following the temperature of a fridge known to be at 12 °C
      While the ambiant temperature readings are fine, the thermocouple readings vary between
      12 deg C and 19 deg C
      I saw on the site that the K meter was requiring a firmware update. I proceeded with the update bu there was no improvment. I also tested the program on M5 Core and M5 Go with the same results.
      Does somebody have some experience with K meter ?

      1 Reply Last reply Reply Quote 0
      • C
        csierra67
        last edited by

        Hi there,

        I am closing this topic. Actually the cause for these excessive variations was the 5V USB adapter I was using.

        1 Reply Last reply Reply Quote 2